atomic absorption troublshouting, Chemistry

Assignment Help:
Dear sir,
We need your consultancy in a technical problem that we are confronting, and we would really appreciate your help
We are in the process of implementing the heavy metal analysis in fish and water by atomic atomic absorption spectroscopy,
I’m still in the sample preparation optimization phase, and I’m confronting some problems,

Il explain in brief the problem in my last results obtained:
For arsenic analysis, 0.2 g sample is digested with Nitric acid At 90?C for 45 min ,then placed in oven for 2 hours at 130 ?C, then diluted to desired volume

1- 0.2 g Unknown fish sample spiked with (2.5ml) 1000ppb arsenic standard, after dilution in 25 ml H2O,the theoretical value must be 100 ppb, but the practical value came out 75 ppb
2- 0.2 g Fish muscle certified reference material ,containing 12.7 mg/kg of arsenic, and since the dilution is in 50 ml, then the theoretical value must be 50.8 ppb, the practical value came out 0.57ppb

Knowing that both samples were treated in the same condition and in the same time, the difference in the end results is very weird, please advise concerning any note that can improve our work,
or detect any mistake has been made,
waiting for your reply,many thanks in advance

Related Discussions:- atomic absorption troublshouting

The maximum number of electrons, The maximum number of electrons that can b...

The maximum number of electrons that can be accommodated in a 3d  subshell is: (1) 2        (2) 10       (3) 6       (4) 14 Ans: 10

Explanation of tar, The tar command was invented to backup directory trees ...

The tar command was invented to backup directory trees to magnetic tape. The name is short for 'tape archive´. A directory tree is a branching structure, b ut a magnetic tape is a

Fajans rule, fajan rule state that large size highly negative charged ion ...

fajan rule state that large size highly negative charged ion form covalent bond

A metal surface is exposed to solar radiations, A metal surface is exposed ...

A metal surface is exposed to solar radiations: (1) The emitted electrons have energy less than a maximum value of energy depending upon frequency of incident radiations

The wave number associated with this light, The frequency of a wave of ligh...

The frequency of a wave of light is 12x 10 14 s -1 .   The wave number associated with this light is: (1) 5x10 -7 m                        (2)  4x 10 -8 cm -1

Conformation energy profile for butane c2-c3 bond rotation, Conformation en...

Conformation energy profile for butane Conformation energy profile for butane C2-C3 bond rotation

CALORIMETRY, when a 0.800g sample of CaCl2 is dissolved in 9.85g of water i...

when a 0.800g sample of CaCl2 is dissolved in 9.85g of water in a microscale calorimeter the temperature of the solution rose from 23.4 degrees celcius to 35.8 degrees celcius.calc

Colloidal particles, a) During coagulation, the colloidal particles stay un...

a) During coagulation, the colloidal particles stay un-aggregated because: b) The aim during water softening is to precipitate calcium and magnesium as calcium carbonate and mag

Write Your Message!

Captcha
Free Assignment Quote

Assured A++ Grade

Get guaranteed satisfaction & time on delivery in every assignment order you paid with us! We ensure premium quality solution document along with free turntin report!

All rights reserved! Copyrights ©2019-2020 ExpertsMind IT Educational Pvt Ltd